[Ri...] Posted June 14, 2022 Share Posted June 14, 2022 I'm not sure what happened or when/how it happened but I cant Reference the Master Sphere and or Qualify the Master Stylus. File: Qualification error -SDO Message -1215.JPG After initial Ref. sphere position. File: Qualification error -SDO Message -1215 (2).JPG Shown in 8 decimal places, yet FAILS. File: Qualification error -SDO Message -1215 (3).JPG Shown in 8 decimal places (same as above) with Limits dialog box shown. File: 8 decimal places (inch).JPG Last attempt, dramatic change in sigma values. Any assistance is appreciated.Qualification error -SDO Message -1215.JPGQualification error -SDO Message -1215 (3).JPGQualification error -SDO Message -1215 (2).JPG8 decimal places (inch).JPG Link to comment Share on other sites More sharing options...
[Ri...] Posted June 14, 2022 Author Share Posted June 14, 2022 I changed the adaptor plate and managed to get the Master Sphere referenced. Although the sigma is high at 0.00029540 (Inch), it should be about 0.000002xxx, which is what it usually comes out to. I am now performing a Fitting position. Then I will re-reference the Master Sphere. Link to comment Share on other sites More sharing options...
[Er...] Posted June 14, 2022 Share Posted June 14, 2022 In Calypso 5 I have had a similar issue. I think what i did was run it in millimeter? Link to comment Share on other sites More sharing options...
[Ri...] Posted June 14, 2022 Author Share Posted June 14, 2022 I started out in metric, then switched to inch. I've noticed faint "whistling" sounds when the CMM makes minute moves. Link to comment Share on other sites More sharing options...
[Ri...] Posted June 14, 2022 Author Share Posted June 14, 2022 Why are my Sigma's for the Master Probe and the Master Sphere all of a sudden, 100% identical on every attempt? Considering its an SDO error, I'm beginning to think it has to do with an SDO database. Link to comment Share on other sites More sharing options...
[Mi...] Posted June 14, 2022 Share Posted June 14, 2022 Is this an XXT? When we start getting this error, that is not fixed by thorough cleaning and controller reboot, it's pointed to a loose, or damaged XXT head. Link to comment Share on other sites More sharing options...
[Ri...] Posted June 14, 2022 Author Share Posted June 14, 2022 Yes XXT. As the stylus is contacting the Sphere (trigger) I can see times when the duration of some points are taking longer than normal, it is noticeable by eye. I cant get past the Master Probe and the Master Sphere having identical sigma's... 👩🏫Typical.JPGNot typical.JPG Link to comment Share on other sites More sharing options...
[Er...] Posted June 14, 2022 Share Posted June 14, 2022 Have you tried Sphere 1? Link to comment Share on other sites More sharing options...
[Ri...] Posted June 14, 2022 Author Share Posted June 14, 2022 There is no Sphere 1, I'd have to create it. Link to comment Share on other sites More sharing options...
[Ri...] Posted June 14, 2022 Author Share Posted June 14, 2022 Is it normal when switching from Inch to Metric (or vise versa) that the Global parameters don't change respectively? Link to comment Share on other sites More sharing options...
[To...] Posted June 14, 2022 Share Posted June 14, 2022 Please sign in to view this quote. Hey Rick, They have always been same values for me. Link to comment Share on other sites More sharing options...
[Ri...] Posted June 14, 2022 Share Posted June 14, 2022 Please sign in to view this quote. They should be. The MasterProbe is the Master. Link to comment Share on other sites More sharing options...
[Ri...] Posted June 14, 2022 Author Share Posted June 14, 2022 (I am now on v7.2.08) The left side of the dialog box is the Stylus The right side is the Master Sphere. In the attached image, there are clearly two separate Sigma's. This is what i typically have seen. And this brings me closer to what is wrong and why I am getting the SDO 1215 error. The Sigma Limit for Inch Mode = 0.0003937 The Sigma Limit for Metric Mode = 0.0100000 It appears when I, at some point, switched from Metric to Inch the Sigma Limit remained at: 0.0100000 Then when I switched back to Metric (I do this all day depending on the B/P) the Sigma Limit became: 0.00000016 Nothing was going to pass Qualification. 🙄 I have Programs that I call Templates. Inch Mode at 4 places, Inch Mode at 6 places and Metric Mode at 5 places. Open a preferred template, Save As, Create Program....Wash - Rinse - Repeat....Typical.JPG Link to comment Share on other sites More sharing options...
[Ri...] Posted June 20, 2022 Author Share Posted June 20, 2022 The work around was to rebuild the Stylus database, it appears to have corrupted itself. Link to comment Share on other sites More sharing options...
[Er...] Posted June 21, 2022 Share Posted June 21, 2022 Please sign in to view this quote. I created a "new machine" to fix ours one time. So I'm wondering "How'd you rebuilt the database?" Link to comment Share on other sites More sharing options...
[Ri...] Posted June 21, 2022 Author Share Posted June 21, 2022 Deleted all the probes, recreated all the probes. Link to comment Share on other sites More sharing options...