[Sh...] Posted April 20, 2022 Share Posted April 20, 2022 We are seeing so issue with repeatablity in the Z axis with this sensor. We had a pretty good crash and now when we calibrate the probes we sometime get an alert that the force has changed since the last calibration. What is ussually involved with replacing this? Link to comment Share on other sites More sharing options...
[Er...] Posted April 21, 2022 Share Posted April 21, 2022 Contact Zeiss about a replacement. We have one busted after 10 years. It took repeated beatings. The busted one has a "rattle." Sometimes we swap TL1 and TL3 XXT from machine to machine. Of course everything has to be requalified. I would take everything apart from the RDS out and use a fresh stylus. It could be the XXT or something is not seated just so. Clean all contacts with 99% isopropyl and look for damage. Link to comment Share on other sites More sharing options...