In my Qualification program I want to prompt the operator to clean or straighten stylus based on the measurement result.

Currently I have messages that come up when the straightness or sigma is OOT prompting the operator to straighten the stylus, but I would like it to automatically rerun these OOT features after the operator confirms they have fixed the stylus.
  • 4 weeks later...
Here is a PCM Example from Calypso user guide :

This should be a good start ..

Task with PCM functions
1.Place a loop with 6 repetitions around the characteristic. 2.Point to the Break Condition field, open the shortcut menu and select Formula... The Formula dialog box appears on the screen.3.In the Boolean Expression (Yes / No Test) field, define “status_5==1” as the abortion condition and click OK.4.Write the following program code into the field for the output parameters of the characteristic: //************************************************************************************************ 
// Measure Circle hole_5 , check roundness; if out of tolerance, interrupt CNC and request next 
// task: repeat feature or measure next feature or terminate CNC run. 
//************************************************************************************************
// Defining starting conditions
status_5 = 2 
hole_5 = getActual("DIN Round_hole_5").actual 
// If out of tolerance, open a window and inquire CNC End, Continue or Repeat 
if hole_5 > 0.01 then
message("DIN Roundness of hole_5 is:" ,hole_5)
status_5 = inquire("1=CNC-End, 2=Continue, 3=Repeat. Enter a number") 
   message(status_5,"Is your entry correct?", "DIN-Roundness of hole_5 is", hole_5) 
      if status_5 == 2 
         then 
         message("CNC run will be continued") 
      status_5 = 2 
      endif 
endif 


if status_5 == 1 then 
   message("The CNC run will be terminated due to excess tolerance in hole_5.") 
   cncBreak() 
endif 


if status_5 == 3 then 
   message("The circle measurement is repeated.") 
      if LOOP1 == 5 then 
         message("CNC run will be terminated due to too many loop runs in hole_5.") 
         cncBreak() 
      endif 
endif 
print("Status is",status_5, "DIN Roundness of hole_5 is", hole_5) 

5.You can now run the measurement plan. 
×
×