[An...] Posted March 16, 2022 Share Posted March 16, 2022 Hello all I've set Clearanc Dat : 0 ,retract distance 0.06 and Clearance group CP+Z (please see attached) Why ...probe is not going inside of connector....?? Thank you in advance0316 zies_.JPG Link to comment Share on other sites More sharing options...
[Pe...] Posted March 16, 2022 Share Posted March 16, 2022 If I had to make a guess... it might be trying to use the wrong stylus in your star system. Open the circle path strategy and make sure you have the correct # selected. Link to comment Share on other sites More sharing options...
[No...] Posted March 17, 2022 Share Posted March 17, 2022 A screenshot of the circle strategy would be helpful. In order to test Perry's suggestion, manually select the stylus that is used in the first circle strategy, set the coordinate display in the traffic light window to base alignment and position the downward stylus over the connector center. The X/Y coordinates should read close to zero. If not, something's wrong. Another possibility is that two styli got swapped at calibration. Test the same way. Link to comment Share on other sites More sharing options...
[An...] Posted March 17, 2022 Author Share Posted March 17, 2022 Omg i did wrong "qualify passive" Thank you all! Link to comment Share on other sites More sharing options...