I have a questions :
I want to put the Square Slot for my element, in this situation a see the all characteristic available for this Square slot and i will see and the position. I don't see the Square slot on ZEISS Calypso option ,only Slot. I attached the characteristic from 2d.
Thank.s

xyzzzzz.pngplot.png

Link to comment
Share on other sites

I don't understand your question much.

If you want to know center position - you can do it with curve. If you want to know width of slot - you can recall points to 3d or 2d lines - they are easiest.
Link to comment
Share on other sites

Thank's again Martin,
You can explain your opinion,method ,for measured this form(is attached). In this moment i measured with "3 d curve" and after i put linear form at this 3 d curve with datum ABC. The problem is the i don't see the nominal value at this from in raport with ABC and i don't see if is correct.
Thx.

xyzzzzz.png

Link to comment
Share on other sites

What are you trying to achieve?

For surface profile you can not use 2d/3d curves - same goes to slots.
But you can create general surface with recall of curves - then you can use surface profile instead of line profile.