VAST Manual Stylus system will not release?


Recommended Posts

  • 9 months later...
I am having this same issue. I am not sure what happened as I came in and there was a note from a second shift employee saying "Sorry" 🤣 The head is a Vast gold I believe. I have tried using manual release/pick up, quick keys to drop the stylus, and tried restarting/shutting down the entire system and rebooting. Nothing is working. Any ideas?
Try going to manual probe change and picking up the probe that is stuck. Does it go thru the weighing procedure as usual ? Make sure you tell the machine which probe is inserted, then check the feed rate over ride knob. Make sure its on. This applies to an RDX head . If it cant rotate because the over ride is set to ZERO it just sits there. No warning, no control of anything!