Is there a way to eliminate the stem/shaft of the ref sphere out of the tensor qualification? I want 180 degree coverage but will crash on ref sphere stem. I was able to do this prior by making degree angle inputs (start ang /360-ang)

Anything available like this? I don't want to reduce sphere coverage.

Thanks!
Link to comment
Share on other sites

Tensor will allways cover full 180° of cal. sphere.

You can mount cal. sphere to another shape to satisfy your needs.
I don't have any of unusual probes, but only one is rotated 45° with knee setted to 45° to handle one corner of a part.
I had to tell Calypso vectors of it's shaft. Otherwise i could only reduce coverage angle.
Link to comment
Share on other sites

Entering the correct shaft vector when measuring the sphere should eliminate the shaft automatically. If not, check the shaft radius in the sphere management.