[Da...] Posted September 22, 2021 Share Posted September 22, 2021 I haven't been on the CMM for a while, but starting back up. Is there a way to tell Calypso to send the results data to a different location, i.e. a network path vs the existing path on the PC? Link to comment Share on other sites More sharing options...
[d6...] Posted September 22, 2021 Share Posted September 22, 2021 Close all programs. Extras > Settings > Environment > Paths > last item Default Path for Results Files. For whatever reason, it doesn't load an "Open" style dialog box. I usually just copy and paste the path I want from Windows Explorer. Link to comment Share on other sites More sharing options...
[Mi...] Posted September 22, 2021 Share Posted September 22, 2021 As stated above. Its global.. You also need to be logged in under the MASTER user to change if memory serves correctly But If you need the results to a certain folder in pdf format... Go to Measurement tab > Multiple reports > Output media > Click the pdf icon for the green check > click radio box with the 3 dots to the right of the pdf icon> click the name of output variable box top right From here you can tell calypso to send the pdf results into a certain folder out on the network. I do it this way for daily checks, incoming inspection etc.. This is all from my OLD memory but should get you where you want to be.... Link to comment Share on other sites More sharing options...
[Da...] Posted September 22, 2021 Author Share Posted September 22, 2021 Exactly what i was looking for. Thanks to both of you!! Link to comment Share on other sites More sharing options...