[Jo...] Posted August 3, 2021 Share Posted August 3, 2021 Hello All, Ran a probe cal on multiple stylus systems after defining cal sphere with masterprobe per usual. Removed ceramic cal sphere from machine. I decided to swap out a single styli after the fact on one of the stylus systems after the entire probe cal. Since the ceramic cal sphere was physically removed from the machine, I have to re-run entire probe cal yes? Not just the single styli? On other non-Zeiss machines and softwares I would have been ok to just calibrate the one styli if I didn't remove the cal sphere from machine. Wondering if the same applies here. Thanks. JM Link to comment Share on other sites More sharing options...
[To...] Posted August 3, 2021 Share Posted August 3, 2021 Load MasterProbe, do Reference Sphere Position and then qualify only changed stylus. However, if changing the stylus could disturb any of the other styli on the same stylus system, then I might consider doing the entire stylus system. Link to comment Share on other sites More sharing options...