Jump to content

6x108 probe hits mast when changing index


---
 Share

Recommended Posts

I have a 2018 Contura with RDS XXT head. When I change probes from my rack that is on the back left of my machine the head indexes and runs into the probe into mast (or at least drags it along the mast) when I have my 6x108 probe on. Is there a global pathing setting I can use to prevent this?
Link to comment
Share on other sites

You'll have to be very careful and test it out, I recommend turning the feed rate to very slow, but in the CNC stylus system change window where it shows your probe rack & probes, if you click the drop down in the middle of the window, there are approach parameters you can adjust. I have a 160mm probe and I believe I set it to 300mm before and after change in the Y+ direction. I'll post pictures soon.
Link to comment
Share on other sites

***Make sure you click on the probe in the rack change window that you want to adjust, the approach parameters will only affect that stylus system***
Link to comment
Share on other sites

Please sign in to view this quote.

Jamie when you say mast, are you talking about the sides of the bridge? Or the mast of the probe rack? If is it the probe rack, do as suggested above. If it's the side of the machine...I'm following this because I have done this more than once. The head doesn't realize that it's going to hit the side of the machine when indexing.
Link to comment
Share on other sites

If it is the bridge, I would think the Outer Clearance Planes are not set correctly.

I was editing someone elses program and they had set the Z- OCP below the granite for some reason. RDS indexed right into the granite as soon as it cleared Y+.
Link to comment
Share on other sites

Please sign in to comment

You will be able to leave a comment after signing in



Sign In
 Share

×
×
  • Create New...