[Da...] Posted January 25, 2021 Share Posted January 25, 2021 Howdy, everybody! My normal Monday Morning Duramax probe qualification hit a snag today. I went to qualify one of the probe systems in the rack, and it said "Qualify Reference Sphere with Master Probe". Well, I figured maybe a power outage or crash over the weekend, so I did a Reference Sphere Qualification with the Master Probe. Sigma values were good, so on to the next probe qualification. Nope. No such luck, again with the "Qualify Reference Sphere with Master Probe". which I had just accomplished. I've done this loop several times now, no change. The Master Probe qualification panel shows today's current date, but somehow that's not being passed on to the other probes. Any hints, tips, solutions? I'll contact the Help Desk, too. I don't trust the qualification data of my production probes if it won't let me check them against the Reference Sphere! Link to comment Share on other sites More sharing options...
[Ri...] Posted January 25, 2021 Share Posted January 25, 2021 Do you have multiple reference spheres? Make sure the MasterProbe is qualified to the correct reference sphere, and that corresponds the same when you go to qualify a different stylus system. Link to comment Share on other sites More sharing options...