Disc probe error
My disc probe gives 0.01mm error after a collusion ( by my colleague). Does any method to bring it back. Visually looking good , I calibrate it using reference sphere .I calibrate it two three times.
How do you go about that Richard? I've pretty much got the same method from a few different people but basically the method I use is to create a short program with an accurate probe like a 3mm Standard, get the size & location of the ring gage and then measure it with the disk probe and tweek those 3 things to match the 3mm Stylus to calibrate it.
You got it.
You create a measurement plan that measures a certified ring gage (the higher the class the better, XX or XXX) with a normal qualified stylus. Then measure the ring gage with your disc probe. You will comp the stylus data manually based on the differences in X, Y, and radius (assuming you are working in the XY plane).