[Za...] Posted December 18, 2020 Author Share Posted December 18, 2020 Just to update on my situation, today marks 2 weeks we've been down. We finally got a Zeiss technician in here yesterday and he did everything he could do and the machine is still erroring out. There were a few things that didn't necessarily cause the issue but were working against us. The CMM itself was not level and this was causing a lot of stress as the machine tried to travel in the -Y direction as it was going uphill. Also, the tech found trace amounts of oil in our air lines (it wasn't a lot, but it's enough that he had to blow it out of the lines). He also re-set the drives and tried a few other tricks with his software, but in the end, nothing seemed to help the issue. We're now to the point of replacing the RDS and/or the RDS adapter plate. The good news is our old XXT might not actually be bad (which is good because we're going to need that $ to go towards the new RDS/adapter plate that we're waiting on now 🤣 ). The technician is here waiting for the part to arrive that was supposed to be shipped next day early am. Hopefully this is resolved very soon... I see a lot of OT in my future 🤣 Link to comment Share on other sites More sharing options...
[Za...] Posted December 22, 2020 Author Share Posted December 22, 2020 Another update to this issue... The RDS and the XXT were both fine, however the RDS Adapter Plate was shot. Once we replaced that and re-calibrated all of our styli, the CMM is functioning great. Link to comment Share on other sites More sharing options...