[Ke...] Posted December 18, 2020 Share Posted December 18, 2020 I have an articulating head on a Contura G2 CMM. Not sure how this got changed, but when I change probes to another certain one, and after it changes, instead of going to -Z(A0,B0) position as it always has and how I want it to, it swings to +X(A0,B90). How do I change this setting so it doesn't rotate coming out of the rack? Thanks Link to comment Share on other sites More sharing options...
[Ün...] Posted December 18, 2020 Share Posted December 18, 2020 Probe calibration comes out of the magazine at the location you last calibrated. Calibrate your probe in the -Z direction, it will be correct. Link to comment Share on other sites More sharing options...
[Ma...] Posted December 18, 2020 Share Posted December 18, 2020 Under the CMM tab go to Stylus System. Calypso will remember the last stylus position before you close this tab. The next time you pick up that stylus system it will automatically move to the last position it was in before the tab was closed. I always make sure the probe is in the down position before closing the tab. Link to comment Share on other sites More sharing options...