Measuring Sphere with different tip angles.


---
 Share

Recommended Posts

I am trying to measure a large I.D. sphere that requires me to come in at different tip angles. I come in at A0B90, scan 90 degrees of arc, A0B-90, 90 degrees, A90B90, 90 degrees, etc...

I performed a test on the calibration sphere by coming in at A0B0 and scanning the sphere. I received (x, y, z, d) = (-.0012, .0005, -.0015, .9843) and a form error of .0001 with sigma = .0000.

Coming in at each of the four angles and scanning 90 degrees of the sphere with each, I received (x, y, z, d) = (-.0004, .0013, -.0011, .9837) and a form error of .0020 with sigma = .0003.

The absolute differences are (.0008, .0008, .0004, .0006).

Is this proof that the four tip angle measurement is not as accurate as the single measurement?

I need to probe a large I.D. sphere and hold concentricity of .001. This doesn't seem possible anymore. Perhaps a star is my only option?
Link to comment
Share on other sites

Please sign in to view this quote.

The probe angles are separate styli under the same stylus system and they all get calibrated every evening.

I'm not entirely sure how the sphere was manufactured.
Link to comment
Share on other sites

Do you recall all the points into one sphere ? Spheres are a real PITA ! If the sphere is generated in software for a milling operation, then they are usually pretty close to spherical. If they are turned in a lathe then there are several variables that will mess with the spherocity, making things very hard to do. In a turned form the actual radius programmed doesn't always coincide with the diameter. Form errors will show differently on different areas of the sphere.
Link to comment
Share on other sites

I don't recall the points, I simply make many "Circle Paths" using different tip angles. The images are at 100x magnification. That being said, I just ran another tip through the same procedure and the results were just fine. This is a problem with the individual tip and it's angles. Perhaps I need a full scanning calibration on it, or perhaps I should replace the tip.

I'll do a bit more testing after a full calibration and make a decision. Thanks Dave.
Link to comment
Share on other sites

1417_ff7214f485f702711605e86bfac04543.jpg
I have had very good success checking Spherical Radii using 4 circle paths as shown. We have to hold .003" profile and this method has worked very well. Maybe something to try.
Link to comment
Share on other sites

Measuring a single feature with multiple tips on an XXT, especially an RDS, is difficult.

I have done this before multiple times on an active head holding .001" Positions, but it is difficult. One thing that I've done to help me, and maybe it's just a placebo effect, but I will manually qualify the styli, and then let the software go back and automatically qualify the tips for me.

I also have (on super complex probes) setup a probe check measurement plan with the reference sphere, and manually tweaked the X,Y,Z values of the probe to get all of the error out. - This is kind of a last ditch effort for me.
Link to comment
Share on other sites

Thanks for the advice Richard.

My issue turned out to be that somebody had crashed the probe coming straight downward and the shaft had gotten pushed up into the body. I tried tightening and realized I couldn't fit a tool in there. I replaced the tip and now everything seems to be in agreement and repeating.

It's strange that our daily calibration didn't notify me that there was something wrong.
Link to comment