LINE PROFILE
Hello,
I have two questions :
1) I scan this profile (i make the section ,i put 3d curve and afte i put in the characteristic from FORM AND LOCATION the Line _profile with ABC and without ABC. It is ok , is necessary to do something else , mentioning that this is a square slot???( i attached the photos with the drawing and with the scan.
2) In photos (3) the scan not is continue see the zoom , is ok, i can make continue and who?
Thanks in advance
If you select this button it should connect your line scan. You should also select line profile under form and location to report your profile result. You should actually do 2 of them. One for the profile to ABC and another as the profile to itself. Also with this being surface profile, if this is a deep pocket, you will want to do this in several sections of the pocket.
Thanks very much for answer ,for clarify and with your amability if you know if is possible measurement directly at this (Square SLOT) ,the Calypso have the option SQUARE SLOT or it is possible to eliminated the Clearence Plane between
two points (for shorten the measurement time)? thank you
You mean Rectangle? But rectangle will give you a center somehow (I don't use it). The requirement you have on the drawing is profile of a surface, so you need the "walls".