After manually aligning my part and run it in CNC mode, all probes are offset by approx. 1/8" or so, but cannot figure out why.
The part is set vertically and I am using the #5 left probe to scan. The part has more features, I only show the back face.
I realigned it, and the same happened. As shown on the picture, the red circle is the probing path, and is supposed to be concentric to the part. I am not scanning the hex on the picture, the probe "fell off"
What can cause this? and how can I fix it, without rewriting the program. I could, but most interested in finding out the cause.
Thanks in advance

IMG_5010.JPGprobing path is off centered.JPG

Link to comment
https://qualityforum.zeiss.com/topic/14356-alignment-is-off/
Share on other sites

Check for blocking movement planes ( in CNC menu ) if it didn't swap machine table plane ( those checkboxes with circle around them ).

If it's not swaped ( rounded are down as Z- ) then you can correct it by choosing right spatial axis.
If it's swaped, then you got to point where i usually create new program from scratch ( i did not found a solution for correcting it ).

From screen it says, that spatial plane is X+, so it can not be setted for Z origin. Also alignment can not be calculated because of this.
Link to comment
https://qualityforum.zeiss.com/topic/14356-alignment-is-off/#findComment-75122
Share on other sites

Please sign in to view this quote.

I changed the origins, and now it works. I cannot believe I missed the X and Y origins, I had them set wrong.
Changed the X to the back plane, the Y to the ctr of bore, and the Z offset from the back plane, also the clr plane for the planar rotation is changed to Z+, and now all good.🙂