(There is head CMM Vast Gold, )How can I calibrate style with d=0.3 mm(feature)?
I think: Stress (about 30% mH) and Taper angle of Sphere (less 80). I have never done it, who know something about it...
Link to comment
Share on other sites

Please sign in to view this quote.


I used to have problems, but were related to the quality of the probe… no issues with Zeiss probes (0.3 , 0.4 & 0.5)
Link to comment
Share on other sites

Please sign in to view this quote.

Not sure if (less 80) is a typo but using less than 180° is quite common in this scenario especially since you probably have a 30mm reference sphere which will likely cause the stylus shank to get in the way. You'll know it's interfering due to the higher sigma (S) values. I typically reduce the the sphere coverage in 5° increments until the sigma falls to an acceptable value. Of course, this is not possible with passive qualification on an XXT sensor.
Link to comment
Share on other sites

Thanks All. I'm going to use 50mN dinamic probe 25% (when I calibration d=0.5 with 100 coverage of sphera (d=30 mm) and stress 100mN ).
Link to comment
Share on other sites

We have a RDS VAST TL1 head on our Contura G2 and a RDS VAST TL3 head on an ACCURA II.
Until we purchased a 8 MM reference sphere we had a 30 MM reference sphere we used for all our stylus.
We have qualified stylus as small as 0.25 MM and would get good results by changing Sphere Coverage in the Probing system qualification window.
I noticed Tom Oakes posted this was not possible with a passive sensor. My question is why is it not possible?
We done it for 9 years until we got the 8 MM sphere.
Greg Kozera
Link to comment
Share on other sites

Please sign in to view this quote.

The passive qualification uses 180° of the sphere, not matter what your coverage angle is set to. The geometry qualification CAN use less than 180°. However, the bending parameters are based on the scanning portion so if your probe is shanking out, the bending parameters could be skewed. Now, to be fair, I have not tested this in years but I still believe it to be true.
Link to comment
Share on other sites

Thank you for the explanation as to why this does not work with passive sensors. What confuses me is we were able to qualify and use the tiny stylus with good results. We set the Sphere Coverage to 120° for the 0.25 MM stylus using the 30 MM reference sphere and while watching it qualify never saw the machine go beyond the 120°. The R value came out at .129 and S value at .0013. Subsequent qualifications using the 8 MM reference sphere yielded near identical results.
Link to comment
Share on other sites

Yes sir. We started at the 180° and kept dropping the angle until we hit 120° as the working angle.
I must tell you Tom you have been a great source for answers to a number of questions I have had over the years, and I need to thank you for that.
Greg Kozera