[Jo...] Posted July 7, 2020 Share Posted July 7, 2020 This morning I turned my Contura on with the control box. Turned the drives on. started calypso and ran the home sequence. the CMM is traveling up in Z all the way to the head and faulting out for limit switch (or over travel). I have tried multiple different things but I keep getting the same outcome. I even tried shift > down return > f12 (to do the manual home) with calypso closed and I'm still getting over-travel in Z... any suggestions or comments on something I may be overlooking? ❓ Link to comment Share on other sites More sharing options...
[Jo...] Posted July 7, 2020 Author Share Posted July 7, 2020 Fixed** I got in touch with Zeiss tech. support and they helped me. It was the glass scale on the side. Had a tiny bit of dust on the bar code at the bottom. which was blocking communication for the limit switch. Link to comment Share on other sites More sharing options...