Jump to content

Qualifying new stylus system with short probes


---
 Share

Recommended Posts

I'm wondering how to qualify this new stylus system(containing front and back styli) without shanking on the ref. sphere when checking the bottom.
Link to comment
Share on other sites

  • 2 weeks later...
And how did he/she determine the exact diameter of that ruby sphere, which is the basis for the whole calibration? 😕

With a known reference sphere you could modify the "opening angle" in the stylus settings in order to avoid collisions, but fur such a short stylus this is no option, as it also narrows down the sector of the ruby that actually gets calibrated.
Link to comment
Share on other sites

  • 3 weeks later...
To get you through a pinch, you can reduce the Ref Sphere coverage from 180 degrees to something less. I normally reduce the coverage until my sigma values are in line with what I am expecting. Obviously, the more you reduce, the less accurate the qualification becomes. This assumes you have an active head.

Just noticed Norbert had the same suggestion.
Link to comment
Share on other sites

  • 11 months later...
I have a VAST RDS-XXT TL3 as passive head. I can't find the way in order to reduce the sphere couverage when I try to qualify my Pointer Stylus. (0.1mm)
If you have some indication or instruction, please I will be thankfull.
Link to comment
Share on other sites

I have tried this option, but ever the path around the Reference Sphere is the same, with false contacts. I have changed the Taper Angle but with Passive Qualification, the unique for my RDS-CAA, the 180 degrees are taken by default, I do not know why.
Here I show the real problem with this Pointer Stylus (626113-5000-004) Zeiss, the same result with other small Reference Sphere.
What to do? Re-qualification considers this Angle? 3108_12d03a522017dd50e01c0bc5231fb8d6.jpg
Link to comment
Share on other sites

 Share

×
×