[Si...] Posted March 26, 2020 Share Posted March 26, 2020 Tilt and Rotation? I have 180 Tilt and 0 Rotate. If I check it at the default 135 degree I always get an error that the MP will smash into the shaft radius. I had a Zeiss Tech verify the size and setup of the 8 MM Ref Sphere when he was here doing cal last month. I figured this setup would be easier?Sphere.jpg Link to comment Share on other sites More sharing options...
[Ri...] Posted March 26, 2020 Share Posted March 26, 2020 I vary rarely keep my RS at 135. It's either at 180 or 90. Make sure when you build the new RS, you select that it is an RSH model, and the correct diameter sphere. Link to comment Share on other sites More sharing options...
[Si...] Posted March 26, 2020 Author Share Posted March 26, 2020 Ya, I used the setup choices that are in the system and the Tech verified it for me. I just want to make sure my Rotation is correct. Link to comment Share on other sites More sharing options...
[Ri...] Posted March 26, 2020 Share Posted March 26, 2020 If the tilt is as 180, then the rotation doesn't matter. It's only when it deviates from 180 that it matters. If you kept your RSH in that same position, but put the tilt at 135, the rotation would be 0. Thanks. Link to comment Share on other sites More sharing options...
[Si...] Posted March 26, 2020 Author Share Posted March 26, 2020 Thanks. I was able to successfully Calibrate my 0.8MM Probe. I must have been having faulty calibration with the Default Ref Sphere. Link to comment Share on other sites More sharing options...