[Ke...] Posted March 5, 2020 Share Posted March 5, 2020 Good Afternoon I stopped the machine during a probe change and of course it doesn't register a probe. So I did a manual pick up stylus system and when I a stylus back in the little wrench symbol came up however it has bee doing it for over 10 minutes regardless of what I do. Anyone else have this happen and know of a possible solution aside from rebooting? Link to comment Share on other sites More sharing options...
[Cl...] Posted March 5, 2020 Share Posted March 5, 2020 Make sure the contacts are clean, and the adapter plate is seated properly. You can "trick" Calypso, by choosing the stylus that you are inserting from the drop-down then manually install it. Link to comment Share on other sites More sharing options...
[Ke...] Posted March 5, 2020 Author Share Posted March 5, 2020 Please sign in to view this quote. Unfortunately right now the software is hung up on the wrench icon. Even as I type here the icon has taken over my normal cursor. Link to comment Share on other sites More sharing options...
[Ow...] Posted March 5, 2020 Share Posted March 5, 2020 Short of re-booting, You can try holding down the Ctrl key and selecting Y . That will bring up the process monitor or an error report that will stop it. Link to comment Share on other sites More sharing options...
[Cl...] Posted March 5, 2020 Share Posted March 5, 2020 Also, on the control box you can push and hold Shift & Lower return simultaneously for a second, then push the "X" button. Link to comment Share on other sites More sharing options...