[Cl...] Posted January 13, 2020 Share Posted January 13, 2020 While moving to store a probe, the adapter plate just fell off the XXT. Any ides why this would happen? Link to comment Share on other sites More sharing options...
[Ky...] Posted January 13, 2020 Share Posted January 13, 2020 Check for dirt between the magnets. And maybe the probe wasn't seated in the pocket properly. We have that exact head and have never had that problem. If you find out post it so I can look out for the problem. Link to comment Share on other sites More sharing options...
[Cl...] Posted January 13, 2020 Author Share Posted January 13, 2020 Thanks Kyle. Turns out the stylus had recently been changed to a slightly longer one, and when it went store the stylus, the tip hit the black spring loaded cover which knocked the probe off. Happened so fast we actually didn't see it hit. Link to comment Share on other sites More sharing options...