XXT RDSCAA Check box Missing and Multiple Control Message


---
 Share

Recommended Posts

After quite a lot of time on Calypso and last year with 6.4 (Qstat interface 1.97), decided to use our XXT for some parts with various angled faces and cylinders. During qualifications, going by great training pamphlet by R.S., making great progress.
However, where I should be checking the RDSCAA checkbox in Stylus System window, there is no check box. Have shutdown and re-zeroed a couple times but still no box. Am I missing something does newer version not have the box. Also get a window when first start calypso that asks, "Do you rwelly want to use more than one control?". Always choose No for this, but am curious as to what's causing it. 😐
Link to comment
Share on other sites

Are you creating new styli? As far as I know, you need to set CAA on a stylus when you first create it. You can not add it to an existing stylus. When you go to create the new stylus system, the RDSCAA box is seen in the image below.
1132_032c404ae5d9e85f2ba4796ac184aa24.png
Is the box missing on your machine?
Link to comment
Share on other sites

Yes, am creating new stylus system. First one seemed to Qualify and work OK for measuring small plastic part. Have installed and am fixing to qualify another stylus. This time 3x50 and also am missing checkbox here also.
Link to comment
Share on other sites