[K...] Posted February 17, 2019 Share Posted February 17, 2019 I developed the 5 mm stylus. However, it won't qualify as its radius is 2.388 after calibration. Other styluses are calibrated and working fine. The Stylus Ball diameter is exactly 5 mm by micrometer. What may be the causes of this. Thank you Link to comment Share on other sites More sharing options...
[Da...] Posted February 18, 2019 Share Posted February 18, 2019 Could it be that the ball is damaged? Or it has somekind of build up on it? Check under microscope. Link to comment Share on other sites More sharing options...
[An...] Posted February 18, 2019 Share Posted February 18, 2019 See attached.Pic_1.pdf Link to comment Share on other sites More sharing options...
[Aa...] Posted February 18, 2019 Share Posted February 18, 2019 What qualification routine are you using? Tensor qualification routines should correct for a reasonable amount of shank deflection. (Unless the shank is made of rubber, like in Andreas' sketch. 🤣 ) Link to comment Share on other sites More sharing options...
[To...] Posted February 18, 2019 Share Posted February 18, 2019 If this is an XXT sensor, make sure there isn't anything stuck to the magnet, the balls are clean, the inside of the XXT sensor is clear of debris. Check to see of the ruby is loose on the shaft. Check to see if probe it tightened to adapter plate. That's all I have. Link to comment Share on other sites More sharing options...
[Da...] Posted October 25, 2019 Share Posted October 25, 2019 Please sign in to view this quote. Bump. Link to comment Share on other sites More sharing options...