[Pa...] Posted May 29, 2019 Share Posted May 29, 2019 Hi, Does anybody know how I could put a condition associated with the master probe calibration i.e. if sigma is greater than 0.002 then re-run the master probe calibration until less than 0.002? Thanks, Paul. Link to comment Share on other sites More sharing options...
[Ma...] Posted May 30, 2019 Share Posted May 30, 2019 Load the MasterProbe. Click on the CMM Tab. In the middle left of the dialog box you will see "Set Limit Values" here is where you set the values. Click on the Define limit values for this stylus and the information on this page can now be edited. Mark Link to comment Share on other sites More sharing options...
[To...] Posted May 30, 2019 Share Posted May 30, 2019 To add to Mark's post, the "Set Limit Values" in the CMM settings is a global setting for ALL styli. If you want to create a unique limit value for a specific styli , i.e. MasterProbe, load the stylus system, activate the desired stylus and click "Set Limit Values" in the Probing Qualification window. Link to comment Share on other sites More sharing options...
[Pa...] Posted May 31, 2019 Author Share Posted May 31, 2019 That's great, thanks Mark and Tom, much appreciated. Link to comment Share on other sites More sharing options...