Hi,
Does anybody know how I could put a condition associated with the master probe calibration i.e. if sigma is greater than 0.002 then re-run the master probe calibration until less than 0.002?
Thanks,
Paul.
Link to comment
Share on other sites

Load the MasterProbe. Click on the CMM Tab. In the middle left of the dialog box you will see "Set Limit Values" here is where you set the values. Click on the Define limit values for this stylus and the information on this page can now be edited.

Mark
Link to comment
Share on other sites

To add to Mark's post, the "Set Limit Values" in the CMM settings is a global setting for ALL styli. If you want to create a unique limit value for a specific styli , i.e. MasterProbe, load the stylus system, activate the desired stylus and click "Set Limit Values" in the Probing Qualification window.