[Ke...] Posted May 7, 2019 Share Posted May 7, 2019 Hello I was wondering if anyone would know why I have 1 probe setup that will not connect when I do an auto pickup from the probe rack. I have 4 probe setups and all the other are fine. I even used it earlier today when I went to pick it up again it would not connect to the rollers and I got an error message. Thank you Link to comment Share on other sites More sharing options...
[Fa...] Posted May 7, 2019 Share Posted May 7, 2019 hi, do u use an Vast XXT TL1? TL3? or an Vast? Link to comment Share on other sites More sharing options...
[No...] Posted May 13, 2019 Share Posted May 13, 2019 What was the error message? The two most common problems (VAST) are: - probe doesn't connect to all 3 rollers (error will say so) --> clean rollers - IDE chip PCB doesn't make proper contact with the two spring-loaded pins in the head --> clean pins and pads on PCB with alcohol. If pads are already corroded (small dark dot visible in center of pads) clean with a glass fibre pen. Link to comment Share on other sites More sharing options...
[Da...] Posted May 13, 2019 Share Posted May 13, 2019 Are you able to pick up the same probe manually? Link to comment Share on other sites More sharing options...
[Ke...] Posted May 13, 2019 Author Share Posted May 13, 2019 I was unable to pick up the probe manually and I cleaned the probe and all worked fine after I did that. Link to comment Share on other sites More sharing options...