[Cl...] Posted March 6, 2019 Share Posted March 6, 2019 My Wednesday hasn't started off on the right foot. Problem #2 for the day, during a stylus exchange, the probe stops above the socket just after dropping off the stylus system, and the attached message appears. Link to comment https://qualityforum.zeiss.com/topic/19102-probe-socket/ Share on other sites More sharing options...
[Ow...] Posted March 6, 2019 Share Posted March 6, 2019 Probably the same thing you're thinking but, I think you've got something wrong with that vast-gold head. Link to comment https://qualityforum.zeiss.com/topic/19102-probe-socket/#findComment-107464 Share on other sites More sharing options...
[Me...] Posted March 7, 2019 Share Posted March 7, 2019 I had the same problem...I relocated my probe rack location and problem was solved. Minor variations in the racks location from a loose bolt holding it to the table etc. caused me this very problem. Link to comment https://qualityforum.zeiss.com/topic/19102-probe-socket/#findComment-107466 Share on other sites More sharing options...
[Cl...] Posted March 7, 2019 Author Share Posted March 7, 2019 Yes, that seemed to eliminate the problem. Link to comment https://qualityforum.zeiss.com/topic/19102-probe-socket/#findComment-107468 Share on other sites More sharing options...