calculation methods for curve
Have anyone know about the calculation method of deviations for the curve. 3D curve has several methods for calculate the deviations. The cookbook explaned following methods. but I am not sure I have actually understand it.
– ➤ Nominal Vector Direction [⇨ 1-67]
– ➤ Act -> Nom [⇨ 1-67]
– ➤ Nominal in plane [⇨ 1-68] (only for 3D curves)
– ➤ Actual in plane [⇨ 1-68] (only for 3D curves)
– ➤ in X direction [⇨ 1-68] (only for 3D curves)
– ➤ in Y direction [⇨ 1-68] (only for 3D curves)
– ➤ in Z direction [⇨ 1-68] (only for 3D curves)
here, i am interested in the "Nominal in plane" method. 919_b247825fb2f7e75116ea291f387b0c0c.png
for the picture above, all points of this curve are at the section Z=3.644, but the Z component of vector are not 0.
The default calculate method is Nominal vector direction, after i sacn it, the actual points is not actully at the Z=3.644 but more or less a bit, the point 1 is 3.644079,other points are 3.6441 or 3.6439 etc... .
I think because the surface is a bit minus material, so the probe go acrossed the nominal point, so the Z value of the actual point is a bit bigger than 3.644, showing in the picture below. 919_3f133035167683a609b3064e5e2c0006.png
So, how can I get the point actually on the Z=3.644?, at "point" feature, it can get by the "net point" space point mode.
I think it can not be probe actually.
But it maybe get by mathmatic calculation, I found the "Nominal in plane" calculate method, when I toggled to this calculate method, the Z value of all the actual points are changed to 3.644. it seems realized what I want, but I am not sure.
below is the explaination about the nominal in plane i found on cookbook . 919_d954065535b47780b316b86f6a0cb3e8.png
but what is the intersection plane? i never defined it. and how the actual point project onto the intersection plane?
Thanks in advance.
https://qualityforum.zeiss.com/topic/19806-calculation-methods-for-curve/
The intersection plane is the plane that connects all the points of the curve. I.e. if you create your curve from a plane-section that plan is the intersection plane.
If you create a 3d curve where not all nominal points share a plane you will not be allowed to use the nominal in plane option.