Qualifying various Stylus Systems for Vast XXT


---
 Share

Recommended Posts

It's been a very long time since I've programmed with an XXT and I'm super rusty with it. We have an RDS-CAA.

What is the suggested method for qualifying these various systems in regard to them being either Sensitive/Standard/Robust with the % as well.

1.5x50mm Star
3x50MM
4x85MM
1x44MM
1.5x85MM
0.5x75MM
Link to comment
Share on other sites

Is your 0.5mm actually have an ML of 75mm or are you factoring an extension? A standard 0.5 x 20 is the only one I would consider reducing the Probing Force and Probing Dynamic for. Me personally, I would use Sensitive and 30% but everyone is likely to have a different opinion on this. These settings work well for me and I don't break a lot of probes.
Link to comment
Share on other sites

For the 0.5x75MM I'm using Sensitive/50% with a 150° Sphere Coverage. It's calibrating .0001mm - .0002mm with 5 positions.

Yes I have long extension attached as well Tom.
Link to comment
Share on other sites

I don't know if anything has changed but during Passive Qualification, sphere coverage is always 180° regardless of what value you enter. Sphere coverage is effective during Geometry Re-qualification.
Link to comment
Share on other sites

Shawn,
Taking the weight limits into the equation, try putting another probe of equal weight on the other side to see if qualification improves.
The probe is probably to small to use a carbon stem?
Even though I was once told by a Zeiss rep that balance didn't matter if the length was within the limits, I can tell you from experience that even on Vast Gold heads, balance plays a critical role.

POST EDIT - I JUST READ YOUR USING AN RDS (BLONDE MOMENT) SO BALANCE, I don't know about with it.
Link to comment
Share on other sites

Jason, what he's using is a probe with a pilot stem that goes into the drilled hole of the ruby.
When you get small ruby's, that is the best method to use when making them. Not all stylus manufactures use the pilot, they just glue the ball to the end and they eventually fall off.
Link to comment
Share on other sites

Please sign in to view this quote.

Sphere coverage is always 180° only for the scanning portion (bending calculation). It will still use what value you have entered when calculating the geometry (6 points at the end) - which is the same thing you will see with a Geometry Re-qualification.
Link to comment
Share on other sites

Please sign in to view this quote.

I would say that weight and balance were the two most important things. The balance can be somewhat forgiving, but the weight is crucial.
Link to comment
Share on other sites

Please sign in to view this quote.

However, the reason some people reduce the sphere coverage is because they're shanking out on the reference sphere. Considering the bending parameters are established during the scanning phase of the qualification, are you not influencing the bending parameters by scanning on a portion of the probe that is more rigid than the actual shank? How much difference would there be in the bending parameters if you didn't shank out because you were using a smaller reference sphere?
Link to comment
Share on other sites

Please sign in to view this quote.

I agree 100%, and the honest question is I'm not for sure. I would have to do side-by-side comparison test to see if there is an effect. My suspicion is that you shouldn't (keyword shouldn't) see much of an issue in accuracy, but more of a performance affect. I suspect that you could potentially see more air-scanning errors.

The official answer is to use the appropriate size qualification sphere when using small probes. I would suggest that if you are using small probes frequently to just swap to the smaller qualification sphere across the board.

Reference sphere, RS-D8-C for RSH-system; Item No.: 600332-8443-000; List price is $450
Link to comment