There is a microchip in the adapter plate. It's designed to hold the name of your stylus system to prevent the accidental mis-loading of a stylus system. I suppose you could have a defective one, but it may just need to have the id re-taught to it.

Go ahead and mount the stylus system in the XXT sensor. Go into the probing system qualification window. At the upper left corner, click the icon for Stylus System Management. In the left hand side, select the stylus system that is mounted and then click the Teach ID over on the lower right side. That should do it, unless there is something wrong with the adapter plate.

You can also turn the ID Chip off if you don't want to use it. Go to the CMM settins window, select Probe on the right, un-tick ID Chip at left.

Tom Oakes | 03-23-2018 01:23 PM |
Link to comment
https://qualityforum.zeiss.com/topic/20900-id-chip/#findComment-118876
Share on other sites

  • 1 month later...

Please sign in to view this quote.

I just tested this on my machine. At the home page, I selected Change Settings > CMM > Probe. Then, I un-checked Activate under ID Chip Recognition. Then, I clicked APPLY. Then, OK. Shut everything down. Re-started the CMM and the checkbox was still un-checked.