Clearance and Retract Issue
I am having an issue with the probe coming in to measure a plane and it is contacting the adjacent plane on entry. Is there a way to shift the probe over in X then shift back to measure. It is a very tight spot and I may need a smaller probe.
Charles Livingston | 04-13-2018 03:21 PM |
Set the clearance dist. to zero and adjust the retract. I personally like using position points, but if its too tight then a smaller probe might be the only way.
Joseph Giacona | 04-13-2018 04:08 PM |
You can also use points that are clear of the adjacent face first. Then check as close as you want. Put another point clear of the adj face as a final point. Since it won't be retracting between points you shouldn't make contact.
Dave Scott! | 04-16-2018 08:00 PM |
I haven't used clearance distance set to anything except zero in 10 years. Prefer to use position points for additional navigation when needed. That's probably what you are seeking here.
Kai KendallKKendall | 04-18-2018 09:58 PM |
I would agree with the above statements. Set your default Clearance Distance to 0 (probably causing the issue) --> Extras>Settings>Measurement. 99% of the time you can fix navigation issues with Clearance Planes, Sub-Clearance Planes, and Retract Distance. I would recommend staying away from CMM position points because you can cause collisions easily if you adjust your measurement strategy. 1% of the time you will need to use Clearance Distance or CMM position. (It is always good to keep in mind that the direction of travel for a Clearance Distance will be in the direction of the feature. Prime example of this is a cylinder that is not perpendicular to a plane, thus you can input a Clearance Distance to navigate an angle stylus into the cylinder.) Hope this was helpful!