You can also use points that are clear of the adjacent face first. Then check as close as you want. Put another point clear of the adj face as a final point. Since it won't be retracting between points you shouldn't make contact.

Dave Scott! | 04-16-2018 08:00 PM |
I would agree with the above statements. Set your default Clearance Distance to 0 (probably causing the issue) --> Extras>Settings>Measurement. 99% of the time you can fix navigation issues with Clearance Planes, Sub-Clearance Planes, and Retract Distance. I would recommend staying away from CMM position points because you can cause collisions easily if you adjust your measurement strategy. 1% of the time you will need to use Clearance Distance or CMM position. (It is always good to keep in mind that the direction of travel for a Clearance Distance will be in the direction of the feature. Prime example of this is a cylinder that is not perpendicular to a plane, thus you can input a Clearance Distance to navigate an angle stylus into the cylinder.) Hope this was helpful!
×
×