[Lu...] Posted Thursday at 05:15 PM Share Posted Thursday at 05:15 PM Hello, Does anyone have a work instruction on how to create an automatic calibration program? Sorry, I am pretty new to Calypso. We have some probes that are labeled RDS CAA. When I qualify these probes, I click the Qualify Stylus button and then take a single hit on the qualification sphere. Then, Calypso automatically calibrates 16 angles. Is there any easy way to reproduce this automatically? Let's say I have a rack filled with 6 different probes. I want all 6 to run at nighttime. Thanks! Link to comment Share on other sites More sharing options...
[De...] Posted Thursday at 06:05 PM Share Posted Thursday at 06:05 PM (edited) Good afternoon, Luke. What you are looking for is found under the "Resources" ribbon. Resources > Utilities > Probing System Qualifications. When running, be sure to specify what characteristics to run. If you have probes assigned to qualifications in your characteristics tab that aren't actively loaded, Calypso will stop and wait for you to manually insert said probe, thus stalling the program all night. For further record keeping, and documentation you can play around with the "Multiple Report" options as well and export to an Excell workbook to reference past calibration numbers and trends. I am currently using the 2019 suite of Calypso, but I would assume the function is located in the same directory. Edited Thursday at 06:06 PM Link to comment Share on other sites More sharing options...