Hi, I normally use stylii with a DK of 1.00mm and above but I have a job coming up where I will require a DK of 0.5mm . I currently use a 25mm reference sphere. My question is will this reference sphere suffice or do i require a smaller sphere for a stylus this small, thanks in advance for any suggestions

Link to comment
Share on other sites

It depends on the MLE of this stylus. If it's 2.5mm, you'll probably need a smaller reference sphere.  I'm not exactly sure where the cutoff is.  I feel like anything above 5mm will be no problem.  I was able to qualify most 0.5 probes with my 25mm ref sphere.  I also noticed on one brand, not Zeiss, that there was a square shoulder between the ML and MLE, which cause the stylus to shank out.  Most styli have a taper between the 2 shanks which gave the necessary clearance.

 

Link to comment
Share on other sites

Please sign in to view this quote.

You need the 8mm sphere for 1mm and smaller styli to qualify 180° of the probe. 