[Wg...] Posted January 23 Share Posted January 23 I have a Sprectrum III machine with RDS-CAA head. When I try to do fitting position with the masterprobe, the retract/approach distance is so much that the machine hits the overtravel limits. Same thing when I manually create the different angles, the probe goes far from the probeing surface. Is there a option that is not well configuered? Link to comment Share on other sites More sharing options...
[Ke...] Posted January 23 Share Posted January 23 Is this a first time setup, or a change in behavior? What is the Radius of your Reference Sphere listed as in the Probing System Qualification window? Link to comment Share on other sites More sharing options...
[Wg...] Posted January 23 Author Share Posted January 23 I recently started in inspection and they never used the RCA-CAA option before. The sphere radius is 0.5906in (15.00124mm) Everything works as it should with a star probe. Link to comment Share on other sites More sharing options...
[Wg...] Posted February 17 Author Share Posted February 17 I needed to reset the MasterProbe to fix this issue. Link to comment Share on other sites More sharing options...
[Ky...] Posted February 21 Share Posted February 21 I know I am late to this particular conversation, but it sounds like it thought that the MasterProbe had a very large radius. You can reset it, like you did, or otherwise just go into the stylus system management window and change the radius to "0" manually. Link to comment Share on other sites More sharing options...