[al...] Posted February 12 Share Posted February 12 Hello I have a T-SCAN HAWAK 2 I calibrate the scanner every morning And the accuracy of the scanner is good in the morning but by the evening and the temperature difference is 3 degrees Celsius The device becomes inaccurate and has an error of about .05 MM for about 200 mm Link to comment Share on other sites More sharing options...
[Ri...] Posted February 12 Share Posted February 12 Please sign in to view this quote. If you have large temperature fluctuations you should perform a quick HyperScale Recalibration. The bar is located in the case. This is the intent of this process. It's extremely quick too. Link to comment Share on other sites More sharing options...
[al...] Posted February 12 Author Share Posted February 12 Thank you. How many degrees of temperature changes do you think are OK? Link to comment Share on other sites More sharing options...
[Ri...] Posted February 12 Share Posted February 12 Hard to say because I don't think the tech specs list anything like a CMM does for allowable temperature change. I would just say that it's probably just a good idea to do the hypserscale recalibration before each scan - I believe this can even be added to your template if you'd like. Link to comment Share on other sites More sharing options...
[al...] Posted February 13 Author Share Posted February 13 Thank you for your guidance. Link to comment Share on other sites More sharing options...