[Pe...] Posted February 24, 2025 Share Posted February 24, 2025 Hi I am currently trying to map Machine No ($K10) with a formula as some of the gauge inputs are varying. Most of the equipment inputs are ok , but one of the vendors is sending $K10 as text rather than numerical which copies over a number with 6 spaces. For example Machine No is "1420", but comes into auto importer as "1420 ", this means the machine catalog is unable to match. I have attempted the following script which is not working, I do not have a manual for this I am not sure if one is available? {if String.Length($K10) = 10 then String.Substring($K10,0,4) else $K10} - I am attempting to check if the length=10chrs and if so trims from the start of the string to 4 chrs. I would appreciate any advice KR Pete Link to comment Share on other sites More sharing options...
[An...] Posted February 24, 2025 Share Posted February 24, 2025 Hi Pete, I have a couple of thoughts. The first is that you could try String.Trim instead of String.Substring. If I use the following: ${String.TrimEnd("1240 ", "")} returns "1240" as a string with no trailing white space. You could also convert the String to an Int. ${ToInt("1240 ")} returns 1240 as an integer which cannot have trailing white space. I also believe that is you used one of these, you wouldn't have to do the If Else check. Hopefully one of these helps. Let me know and I'll see if I can find another method that better suits your needs. Best Regards, Andy Block Link to comment Share on other sites More sharing options...