Hello, I have been doing some research into probe qualification programs due to some doubts i have on the qualification program i currently run weekly. My main question is that i noticed whenever i run the program i do a 5 point pickup on the reference sphere as a manual alignment. Could someone tell me if the manual alignment would overwrite the "reference sphere position" i do before i run the qualification program? I am worried that i might be calibrating the probes off of 5 points instead of the "reference sphere position" instead.

Secondary question is that in the instruction sheet i am reading about how to create a probe qualification program it says to not add a probing system qualification characteristic for the masterprobe if the reference sphere gets moved around. Our reference sphere does definitely move around and so i am wondering if that causes any issues whenever the qualification program is ran even though I reference the sphere position before (which also goes back to my first question)

we just finished the yearly calibration on the cmm, and i wanted to make sure when i qualify all the probes i wouldn't be messing everything up. If these are actually problems i will probably have to make a new qualification program. All responses are appreciated, thank you!

Link to comment
Share on other sites

---

Did the instruction sheet tell you to create a base alignment?  Typically, "most" don't create a base alignment.

If your ref sphere moves around a lot, I would leave it out and plan on doing a ref sphere pos manually.  Or, leave it in and only select the systems you want to qualify and run Current Selection.

 

Link to comment
Share on other sites

---

Creating a base alignment wasnt brought up in that instruction sheet, so i was getting worried that it would use the base allignment for the sphere position instead of the "reference sphere position" in the probing system qualification menu

Link to comment
Share on other sites

---

you don´t need any alignment in a qualification program, I would leave it out

like Tom said, you can do a manually sphere qualification before starting the program, then it is equal, if you leave the sphere qualification characteristic in or not