[Ni...] Posted April 7, 2025 Share Posted April 7, 2025 I have an Accura with a VAST XTR head and westartibg having an issue with one of our probes not calibrating within our less than 1 micron stdev. It is a 90 degree 4mm ruby 150mm long. All other probes have no issues so I built a new probe file, replaced probe and adapter plate, but no luck. I have a handful of news probes not qualifying under 15 microns Link to comment Share on other sites More sharing options...
[Ky...] Posted April 8, 2025 Share Posted April 8, 2025 Is the stylus itself 150 mm long, or does this include an extension. Unfortunately, if it is a very long stylus, it might just be too bendy to actually qualify below the 1 micron threshold. I was working with a stylus a couple of months ago now that was a bit smaller and I think a bit shorter (I can't remember exactly how much) and I wasn't able to get it to qualify below like 2 microns. Hopefully it isn't giving you 15 microns. That sounds a bit excessive, but I don't have the stylus in front of me, obviously. If you can, I would try to replace at least some of the length of the stylus with an extension. Extensions tend to be much more rigid have a smaller effect on the qualification, but if not, sometimes you do need to just live with the larger sigma if there is no way around it. It is an increase in your uncertainty, but if you can't measure it any other way, there isn't really much you can do. Link to comment Share on other sites More sharing options...
[Jo...] Posted May 1, 2025 Share Posted May 1, 2025 I had this problem in the past where the probe was too bendy as Kyle says. I turned the probing behavior to sensitive and also turned down the probing dynamic in %. This seemed to help but it was never as accurate as some of my more rigid setups. Link to comment Share on other sites More sharing options...
[Ma...] Posted May 5, 2025 Share Posted May 5, 2025 You can alternativity add a counter weight to help balance it out. Link to comment Share on other sites More sharing options...
[Tu...] Posted September 30, 2025 Share Posted September 30, 2025 I normally calibrate my probes every morning, upon doing that I see that the sigma value is 0.0007mm is that normal? Please advise? Thanks Link to comment Share on other sites More sharing options...
[DW...] Posted October 2, 2025 Share Posted October 2, 2025 Please sign in to view this quote. Please sign in to view this username. Assuming this is not a typo, repeatability in the sub micron range is very good. Link to comment Share on other sites More sharing options...
