Hi all,

Our DS 0.2mm x L 18mm (626103-5122-018) keeps breaking. It has broken during re-qualificaiton at 15mN force and 5% probing dynamic and during parts inspection. While observing the probe during a qualification with a digital microscope, that probe is flexing.

Does someone also have experience with 0.2mm probes that can share some wisdom?

Thank you!

Image_20250415_110625_527.jpeg

Link to comment
Share on other sites

---

I would exclude dynamic tensor - this probe would be better just for points - scanning can be tought for this probe

Link to comment
Share on other sites

---

Hi Martin,

Good point. We did in fact change the qualification mode for dynamic tensor to tensor after breaking it the first time. However, while re-qualifying it, it broke again. 

We'd like to come up with a recipe that allows us to reliably use 0.2 mm probes. Avoiding scans could be a part of the recipe. 

Thank you.

Link to comment
Share on other sites

---

Please sign in to view this quote.

Many probes have some amount of flex in them.  The function of the qualification, i.e. Tensor, Dynamic Tensor, Passive qualification, etc., is to measure the amount of flex that a stylus has, often referred to as the "bending map". Once this has been established, it is used to correctly measure features.

 

 

Link to comment
Share on other sites

---

Hi all, 

Thank you everyone for your inputs. I used the six point calibration and the probe seems to be holding up for now.

Link to comment
Share on other sites

  • 3 months later...

This topic interests me because I have a small 0.65mm internal square slot to measure.  The minimum entrance diameter is most efficiently acquired by vision system by our supplier.  And now I'm trying to replicate their In-spec readings with a 0.5mm probe.  Does anyone have advice as so far I keep getting values way too high.  It appears there is an optimum depth at which to check the opening which I must hunt for.

I'm using a large 12/18/10 Prismo with VAST probe head.  

Based on the behaviour of the CMM when I tried to use Curve to hunt for the narrowest point - a probe broke - I wonder if I am using the wrong instrument, ie. such a massive gantry relative to part size.  Would a machine like the O-Inspect be better suited for such small parts?  I know a salesperson would be more than happy to give a presentation, but I don't think our company will be buying new equipment anytime soon.  

Thanks for any feedback.

Edited
Link to comment
Share on other sites

Please sign in to view this username.

 Using laser or optics you are getting less deep while lowering diameter. So i would expect for 0,65mm to be depth max 1,5mm - but it's just a wild guess.