[To...] Posted April 22, 2025 Share Posted April 22, 2025 RE: New model of Contura I know there are situations that cause the drives to shut off but I want to know if there a log file that records when and why the drives shut off. Link to comment Share on other sites More sharing options...
[Mi...] Posted May 8, 2025 Share Posted May 8, 2025 I little while ago I had some drive shut off issues on our Contura, kept shutting off on Z+ moves. Fixed it by increasing the air regulator to 5 bar. Link to comment Share on other sites More sharing options...
[Ma...] Posted May 8, 2025 Share Posted May 8, 2025 Hehe i got an issue with machine shuts of drives - no clue why. Just sporadic. But only on 6-14h shift. After a week we found, that worker pulled lever to 30% so air pressure was fine until air grinder was turned on 😄 Link to comment Share on other sites More sharing options...
[Ni...] Posted February 18 Share Posted February 18 Please sign in to view this quote. How do you adjust the air regulator? We have a similar issue; the controller is set at 4.9. Link to comment Share on other sites More sharing options...
[To...] Posted 6 hours ago Share Posted 6 hours ago We are currently experiencing the same issue with drives shutting off in Z movement. It looks like are air regulator is already set to 5. Any other information would be helpful. Link to comment Share on other sites More sharing options...