Hello all,

After i imported SSC file, 1# styli is corrected. But when i created 2# or 3#..., it's always not corrected. Can you give me your points?

image.thumb.png.427817ef1bc195876b69668393e1ef3e.png

Edited
Link to comment
Share on other sites

---

Then i checked the data for the styli as following:

image.thumb.png.073c3da4901860d090570d7532182748.png

Link to comment
Share on other sites

---

Hello JinXi,

 

attached you find a manual for the RDS XXT Simulation, its a bit more complicated with the RDS then with a Vast Gold for example.

XXT - Simulation with a RDS-VAST-XXT.docFetching info...

With CALYPSO >2020 you need the following lengths for the Stylussystem Settings (in the manual is the Zs for the older Versions):

image.thumb.png.2d6a0ed6507ffa5a45357ab4e474d3f1.png

 

Best Regards

Marcel

Link to comment
Share on other sites

---

Please sign in to view this quote.

Hi,

Thank you!

I set the Zs 188.1020, but it still not work. 

image.thumb.png.436e202566cbcc72f9fe06d7292ce757.png

image.png.13c6388e6d02e7a6e55a0f0fce452686.pngimage.png.13c6388e6d02e7a6e55a0f0fce452686.png

i set the data is from the info from the master probe.

image.thumb.png.a5f33005a9cbe9373c070d959997760f.png

image.thumb.png.529c69bac25d799845441c26b83ae08e.png

this data is corrected for styli #1, but once build another the styli(#2 ,#3), something is wrong. i don't know what's the wrong.

 

Link to comment
Share on other sites

---

Easy to overlook - Make sure the diameter for your MasterProbe is correct and not set to 0

  • Like! 1
Link to comment
Share on other sites

---

Please sign in to view this quote.

Good point, that has caught me too many times.