[Lu...] Posted May 23, 2025 Share Posted May 23, 2025 We are using a star probe with a Duramax. I set my clearance cube very tight around the part, and when the star probe uses the X & Y clearance planes it uses a much larger cube than what I have set. My clearance distances are set to 0. Is this some issue with star probes? Any way to resolve it? I have to run thousands of parts, so I need my movements as tight as possible. We are palletizing the parts very tight. Link to comment Share on other sites More sharing options...
[Cl...] Posted May 23, 2025 Share Posted May 23, 2025 Is the probe moving in a large arcing motion by chance? Link to comment Share on other sites More sharing options...
[Lu...] Posted May 23, 2025 Author Share Posted May 23, 2025 Please sign in to view this quote. No, it's just acting like the clearance cube is larger. Link to comment Share on other sites More sharing options...
[To...] Posted May 23, 2025 Share Posted May 23, 2025 In the CMM Settings under the Probe tab, there is a diameter value that creates a safety zone around the XXT sensor. If the stylus system diameter (not the ruby) is larger than this value, then the stylus tips will move to the clearance plans, as you probably expecting. If the stylus system diameter is smaller than this value, the sensor will move out to the clearance planes. It's much worse on an XT. Even worse on some clearance planes when using an RDS/XXT. If it's too far for your application, consider using a +Z clearance plane and rely on the clearance distances or position points to navigate safely. Link to comment Share on other sites More sharing options...
[Lu...] Posted May 23, 2025 Author Share Posted May 23, 2025 Please sign in to view this quote. Thanks. That is what I have done (clearance distance and position points). I am not understanding entirely, but what if we were to turn off this safety zone under the CMM settings probe tab? Link to comment Share on other sites More sharing options...
[To...] Posted May 23, 2025 Share Posted May 23, 2025 I've never considered doing this, so I don't know what will happen. Link to comment Share on other sites More sharing options...
[Ke...] Posted May 23, 2025 Share Posted May 23, 2025 In the Extras-Settings-CMM-Probe menu, what is the value in the Diameter window near the top? Link to comment Share on other sites More sharing options...
[Cl...] Posted May 23, 2025 Share Posted May 23, 2025 Not directly related, but there is a similar type of clearance cube for the rotary unit on the O-inspect 543. However, this one you cannot edit. It can cause problems if your program needs to probe too close the rotary. Link to comment Share on other sites More sharing options...
[Lu...] Posted May 23, 2025 Author Share Posted May 23, 2025 I tried changing that setting (Extras-Settings->CMM->Probe->Diameter Max). It was set to 100mm. I changed it to zero, and it fixed my issue. However, I'm not sure what it's supposed to be set to. The foam cover on my Duramax Z-axis measures 52.7mm. So, should I change this setting to 52.7mm (am I interpreting this setting correctly)? I don't want any crashes due to me misinterpreting this setting. Link to comment Share on other sites More sharing options...
[To...] Posted May 24, 2025 Share Posted May 24, 2025 (edited) Please sign in to view this username. Earlier I mentioned trying the +Z clearance plane along with the possible clearance distance adjustments and/or position points. That will totally ignore the safety zone. Did the +Z CP not work for you? Edited May 24, 2025 Link to comment Share on other sites More sharing options...
[Ja...] Posted May 28, 2025 Share Posted May 28, 2025 Please sign in to view this quote. I have a Duramax and mine is set to 100mm. as well. I would assume that is the default value. Link to comment Share on other sites More sharing options...
[Ma...] Posted May 29, 2025 Share Posted May 29, 2025 A simple and safe fix would be to block edges in Plan/ Navigation/ Block Edges. Block the movement around the x and y axis and leave Z open. This will force all CP movement to the Z clearance plane. Link to comment Share on other sites More sharing options...
[Ma...] Posted June 4, 2025 Share Posted June 4, 2025 Hi, as you stated correctly the 100mm is the default value and only should be changed if you know exactly why you are doing it (since its a global setting). It is a safety feature for the "quill" of the Z-Axis! You can activate the "Narrow Navigation Paths" Or use the solution that Please sign in to view this username. mentioned above. Best Regards Marcel Link to comment Share on other sites More sharing options...