[To...] Posted Monday at 06:24 PM Share Posted Monday at 06:24 PM What sigma value should the probe have to ensure its measuring accuracy? I have heard .001 is ok but also heard .0002 is what it should be at or under. Link to comment Share on other sites More sharing options...
[Jo...] Posted Monday at 06:47 PM Share Posted Monday at 06:47 PM (edited) Where im at we aim for .0015 mm or lower, but ive heard many different answers to that question. Edited Monday at 07:02 PM Link to comment Share on other sites More sharing options...
[To...] Posted Monday at 06:57 PM Share Posted Monday at 06:57 PM (edited) add "/25.4" to the formula EDIT someone edited their post...lol Edited Monday at 06:58 PM Link to comment Share on other sites More sharing options...
[Ha...] Posted Monday at 06:59 PM Share Posted Monday at 06:59 PM At Zeiss, our recommendation for the limiting value is anything under 0.001 mm for VAST XT and VAST XXT sensors. One major exception are probes that are qualified in RC-CAA mode for VAST XXT sensors, where the sigma values typically range between 0.002 mm - 0.020 mm. Link to comment Share on other sites More sharing options...
[Ke...] Posted 15 hours ago Share Posted 15 hours ago New probes are typically below 0.001mm (I usually see 0.0003-0.0005mm). I definitely inspect probes when the sigma increases (0.001mm or higher); although it is best practice to visually inspect probes more often, regarless of the recorded sigma values. Link to comment Share on other sites More sharing options...