Strange scans from L probes on RDS XXT Sensor.


---
 Share

Recommended Posts

What could cause an L probe to create nominals like these in the images attached.

Also note, when the L probe is scanning the circle the Y axis oscillates approximately 140microns on one CMM and appears to be worse on another CMM.

The vector direction changes at 180° to the opposite.

The probe tip is not in contact with any other surfaces during the circle scan.

1000001023.jpg

1000001024.jpg

Edited
Link to comment
Share on other sites

Please sign in to view this quote.

This doesn't happen on my CMMs. Particularly the oscillation of the Y axis.

On the affected CMM's (another shop), when the part is aligned to .0004" or less (In the X axis) the Y axis moves +Y .005-.006" and then at 180° of the circles, moves back -Y .005-.006" (Approximately 140 micron)

I performed the same test on my CMM's, and the Y axis movement was nearly unnoticeable on an .0005" indicator.

Whether using a VAAST XT, or RDS XXT I have never seen this before using any type of probe configuration.

Link to comment
Share on other sites

Please sign in to view this quote.

I see it all the time when using a star or horizontal stylus.  Just to be clear, I am referring to the change in the probe vectors. I’m not sure where you are picking up the oscillation values from. 
 

I also verify that I am maintaining at least a 3 to 1 ration of form to sigma. Quite often it is 10 to 1. 

Edited
  • Like! 1
Link to comment
Share on other sites

Please sign in to view this quote.

Odd that I have never seen this and I always pay attention to the touch or scanned points.

The oscillation, I think is due to third party calibration. 

Link to comment