[Co...] Posted July 8, 2025 Share Posted July 8, 2025 I work at a production shop and a lot of our jobs we run every part through a CMM program as they come off the CNC machine. How often should I be running my qualification program? Normally I would only do so if a probe breaks or if I set up a new probe. I'm wondering if I should be qualifying the styli every week now that we are using the CMM all day during production. Thank you! Link to comment https://qualityforum.zeiss.com/topic/23897-how-often-should-i-qualify-styli/ Share on other sites More sharing options...
[Ja...] Posted July 8, 2025 Share Posted July 8, 2025 (edited) Probably lots of trains of thought on this and while this answer is unhelpful it would be - it depends. How clean is your CMM environment, is junk prone to accumulated on your probes? What kind of material are the parts you are running? If it is a rough surface and you are using scan paths then you may wear down the probe tips without realizing and not catch it until your next qualification. Are you the only one operating the CMM or have full confidence in other operators not to have a crash? How would you know if something wasn't out of whack without qualification at regular intervals? That said for us I run a basic qualification program at the beginning of each shift on our CMMs that covers our main couple of stylus systems that cover about 90% of the parts we inspect. The program takes about five minutes to run and gives me some peace of mind that something didn't get screwy if I wasn't around to babysit the CMM or operators running on it. I do a full requalification if a probe is being replaced or setting up a new one, otherwise that daily qualification is sufficient for our needs. We could probably get away with a weekly qualification but it's quick enough to just get it in every morning. Edited July 8, 2025 Link to comment https://qualityforum.zeiss.com/topic/23897-how-often-should-i-qualify-styli/#findComment-134901 Share on other sites More sharing options...
[Ch...] Posted July 9, 2025 Share Posted July 9, 2025 So once a week would be recommended. We used to do it every morning, but then production complained that it was blocking the CMM for too long. Since then, we only do it once a week. Or in case of a major collision. Link to comment https://qualityforum.zeiss.com/topic/23897-how-often-should-i-qualify-styli/#findComment-134905 Share on other sites More sharing options...
[Da...] Posted July 9, 2025 Share Posted July 9, 2025 When I wanted an answer to the exact same question, I chose to try a few things out. Instead of calibrating after a certain time, I regularly checked my calibrations for two things: 1. Does the probe still measures the correct diameter of the ref sphere and 2. If I use multiple probes in one program, how far do the X/Y/Z coordinates of the ref sphere spread. I found that a probe holds its diameter calibration value quite stably easily over at least two weeks, but multiple probes tend to drift away from each other after some time. So, what I do is, when I use only a single probe in a program, I check daily if it still holds its diameter value. No calibration needed. But if I use multiple probes in a program, I check them daily to each other and if they aren't within my spec (I want them to hold their position to each other within 5 microns), I calibrate them all in one go (but just the ones that I need). I'm pretty happy with that. 1 Link to comment https://qualityforum.zeiss.com/topic/23897-how-often-should-i-qualify-styli/#findComment-134909 Share on other sites More sharing options...
[Cl...] Posted July 9, 2025 Share Posted July 9, 2025 Try adding a daily master part verification. This will immediately alert you if there is any problem with the styli. Or instead of a master part, create an artifact with multiple geometries. Use it to measure size/form/position/orientation of those geometries using every styli rotation/angle. This may take 10-15 minutes, but is well worth the time. Link to comment https://qualityforum.zeiss.com/topic/23897-how-often-should-i-qualify-styli/#findComment-134919 Share on other sites More sharing options...
[Ow...] Posted July 9, 2025 Share Posted July 9, 2025 As defined above, every work scenario is different. We have a lot of Duramax CMM's that I set the global stylus limit values to require a qualification of all probes every 24 hours, with a sigma value of 0.002mm max. The protocol requires that all styli be cleaned before running the qualification program. I do this because on some of them, several different operators use the CMM (set to run in the autorun interface) on different shifts and there is no way to absolutely ensure every part is clean before every part is run and dirt/debri/sticky-stuff can accumulate on the probe that can not be seen with the naked eye and will affect results. I have a daily (every 24 hours) qualification program set to qualify as passive (doesn't take long) and a once a week (generally Monday) that qualifies with the geometry qualification, which takes about 25 minutes. In the limit values, you can set how many hours elapse before the CMM won't run without the styli being qualified. 2 Link to comment https://qualityforum.zeiss.com/topic/23897-how-often-should-i-qualify-styli/#findComment-134945 Share on other sites More sharing options...
Popular Post [Ke...] Posted July 9, 2025 Popular Post Share Posted July 9, 2025 I will add another point of view, which is that the time between calibrations/probe checks determines how many parts need to be quarantined and sorted if an issue is found. Calibrate Monday, run all week, Calibrate the next Monday and find an issue? 1 week of production to quarantine and sort. Calibrate Monday, run all day, Calibrate Tuesday and find an issue? 1 day of production to quarantine and sort. So calibration frequency may be less about the CMM losing accuracy over a certain period, and more amount what is an acceptable amount of risk for the company 5 Link to comment https://qualityforum.zeiss.com/topic/23897-how-often-should-i-qualify-styli/#findComment-134947 Share on other sites More sharing options...
[Tr...] Posted July 9, 2025 Share Posted July 9, 2025 Please sign in to view this quote. Where do you turn this on at inside of Calypso? I thought I had it turned on and tried lessening the timeframe to where I know they haven't been done since, but it still allowed for the part program to run. Thanks in advance. I'd like to turn it on to where it has to happen within 4 days of the last one. I thought it was here, but I am assuming I need to click the activate on the monitoring criterion? Link to comment https://qualityforum.zeiss.com/topic/23897-how-often-should-i-qualify-styli/#findComment-134977 Share on other sites More sharing options...
[To...] Posted July 10, 2025 Share Posted July 10, 2025 I have heard the following responses in my career as a Zeiss guy. once per shift once per day once per week once per month I'm supposed to qualify my probes? as needed whenever I am doing a part with tight tolerances 3 1 Link to comment https://qualityforum.zeiss.com/topic/23897-how-often-should-i-qualify-styli/#findComment-135043 Share on other sites More sharing options...
[Kh...] Posted July 10, 2025 Share Posted July 10, 2025 Whenever you need a break LoL 2 Link to comment https://qualityforum.zeiss.com/topic/23897-how-often-should-i-qualify-styli/#findComment-135057 Share on other sites More sharing options...
[Tr...] Posted July 10, 2025 Share Posted July 10, 2025 Please sign in to view this username. Recently we put into effect across 8 CMM's we use that 1 time a week is no longer sufficient, and we now require minimally 2x a week. Recommended to do daily. The CMM in my specific lab I'm keeping a 3x a week(Mon, Wed, Fri) consistently since we are saving probe qualification sigma value's reports. But to ensure a consistency want to test putting a limit in the system on it to require it to be done. Hopefully Please sign in to view this username. can provide me what I'm missing to making this efficient from my quoted post above. Link to comment https://qualityforum.zeiss.com/topic/23897-how-often-should-i-qualify-styli/#findComment-135071 Share on other sites More sharing options...
[Ow...] Posted July 10, 2025 Share Posted July 10, 2025 Please sign in to view this quote. Travis, Sorry, I rarely get time to get on here and answer questions anymore, just opinions:-). I have it set-up just like you do, except @ 23 hours (0 in days) with both the sigma and Maximum permissible time check-marked as active. Most of our Duramax's are running version 6.0 to 6.2 and according the operators that get here way before I do, it works. Link to comment https://qualityforum.zeiss.com/topic/23897-how-often-should-i-qualify-styli/#findComment-135077 Share on other sites More sharing options...
[Ch...] Posted July 10, 2025 Share Posted July 10, 2025 also there is ... calibrate probes every time you inspect a part, you can build it into a program, it's a bit overkill, but perhaps needed for tighter tolerances. Link to comment https://qualityforum.zeiss.com/topic/23897-how-often-should-i-qualify-styli/#findComment-135079 Share on other sites More sharing options...
[To...] Posted July 14, 2025 Share Posted July 14, 2025 Please sign in to view this username. If I am not mistaken, isn't the qualification done after the Base Alignment? Link to comment https://qualityforum.zeiss.com/topic/23897-how-often-should-i-qualify-styli/#findComment-135159 Share on other sites More sharing options...
[To...] Posted July 14, 2025 Share Posted July 14, 2025 Does anybody do probe verifications instead of qualifications? 2 Link to comment https://qualityforum.zeiss.com/topic/23897-how-often-should-i-qualify-styli/#findComment-135161 Share on other sites More sharing options...
[Ma...] Posted July 14, 2025 Share Posted July 14, 2025 Please sign in to view this username. I do probe verifications per the Calypso Basic course. This consists of +/- 0.003mm for XYZ on probes using the reference sphere. I do this once a week when I do my mass re-qualification. I have recently had a few questions about this though and maybe you can help me out. We get a lot of parts that have radial runout callouts ( 2 cylinders in the same axis ) and sometimes i need to use 2 different probes to measure the features (probe 3 does datum feature and probe 4 does feature) If my XY values on my verification aren't the same for probe 3 and probe 4, do you know if this can skew the results for runout? I know you can go in and alter the XY values of probes to match one another but I am unsure if this is illegal. 2 Link to comment https://qualityforum.zeiss.com/topic/23897-how-often-should-i-qualify-styli/#findComment-135185 Share on other sites More sharing options...
[Cl...] Posted July 15, 2025 Share Posted July 15, 2025 (edited) As stated, prior. We run master parts daily, documenting results. Which must fall within specified limits. This verifies (among other things) all styli. Edited July 15, 2025 1 Link to comment https://qualityforum.zeiss.com/topic/23897-how-often-should-i-qualify-styli/#findComment-135211 Share on other sites More sharing options...
[Ow...] Posted August 12, 2025 Share Posted August 12, 2025 Please sign in to view this quote. For the record, last week I was able to verify that this qualify every 23 hour mandate does and is working as I defined above. If not qualified in that time frame, the CMM pops up a window that basically says the probe needs qualified. Link to comment https://qualityforum.zeiss.com/topic/23897-how-often-should-i-qualify-styli/#findComment-136707 Share on other sites More sharing options...
[Ri...] Posted August 15, 2025 Share Posted August 15, 2025 Please sign in to view this quote. I have been doing this for almost 18 years now, once a month and I move the reference Sphere around the CMM in 10 locations. 5 at granite level and the same 5 raised 6 to 8 inches. In a 10-month period I will have 20 reports of 5 standard articulations for all sensors we use including multi-Sensor machines. VAST XT, VAST XTR, VAST XXT. This is how I evaluate the volumetric area, sensors and styli. (Excel and graphs, etc.). Link to comment https://qualityforum.zeiss.com/topic/23897-how-often-should-i-qualify-styli/#findComment-136859 Share on other sites More sharing options...