[Lu...] Posted Wednesday at 07:50 PM Share Posted Wednesday at 07:50 PM (edited) Hello, I was doing the "Ref. sphere position" and the first point I took "in the probing direction" showed an error saying that the previous masterprobe weight and the current weight is different, and is asking me if it should be "redefined". It goes on to tell me that it could be from adding a probe to the masterprobe, and a couple of other options none of which i had done. Could this mean that the CMM was damaged? I am worried that if I redefine the weight and qualify all of my styli that it could cause more issues in case the CMM was actually damaged and the weight sensors were incorrect. Could air pressure also be a potential cause? Any advice is appreciated. The machine is a Contura G2 with a vast xxt rotating probe head Edited Wednesday at 07:56 PM added probe head info Link to comment Share on other sites More sharing options...
[Cl...] Posted Wednesday at 08:27 PM Share Posted Wednesday at 08:27 PM I get the same error intermittently on a couple of my Conturas. I clear it and don't have any further problems. 1 Link to comment Share on other sites More sharing options...
[Ky...] Posted Wednesday at 08:37 PM Share Posted Wednesday at 08:37 PM If this is consistent, Luke, then I would call your HW support team. There is a spring inside of the XXT that weighs the stylus system and if that gets stretched or otherwise damaged, it can cause that behavior. If it only happens once, I wouldn't worry about it too much. Likely some sort of fluke. I would make note of it, but wouldn't think too much of it until I see a consistent history of it and especially if it is getting worse. 1 Link to comment Share on other sites More sharing options...
[Lu...] Posted Wednesday at 08:54 PM Author Share Posted Wednesday at 08:54 PM Please sign in to view this quote. Forgot to mention, it says it is heavier by "0.012d mm". Would that be ok to calibrate it? I just want to make absolutely sure before I mess anything up lol thank you Link to comment Share on other sites More sharing options...
[Ky...] Posted Wednesday at 10:43 PM Share Posted Wednesday at 10:43 PM You aren't going to mess anything up that isn't already messed up. If it is an issue with the XXT, you'll have to requalify everything anyway. 1 Link to comment Share on other sites More sharing options...
[Ri...] Posted Thursday at 11:40 AM Share Posted Thursday at 11:40 AM I found when I get this, I clean the adaptor and contacts, then the weight goes back to normal. 1 Link to comment Share on other sites More sharing options...