[Pe...] Posted 4 hours ago Share Posted 4 hours ago New problem with a contura, wonding if anyone has seen. Just started happening, only thing new that happened is that IT changed the battery in the UPS. On startup there is a very generic error message: "wbScanLink-Error! Maybe its not installed correctly." After startup and homing, calypso doesnt recognize that there is a probe in the sensor. When I tell it to pick it up, it asks me to create a new stylus stystem out of a laser scanner. But, this is a RDS XXT with TL3 sensor, so I dont know why this be happening. Link to comment Share on other sites More sharing options...
[Pe...] Posted 2 hours ago Author Share Posted 2 hours ago I'll consider this fixed. We happened to have a zeiss tech here something else whom was able to provide us with a solution. For anyone that might come across the same issue: The wbScanLink-Error was cause by somehow an additional sensor being checked in the sensor list. Linescan as well as the Renishaw TPx were checked despite us not having them: The other issue with the laser scanner was that somehow Calypso was connected to one machine, the ConturaG2_RDS, but there was a second machine profile that despite not being connected to, had the OSIS laserscanner set up and somehow it was drawing from that. We deleted that second machine and it went back to normal. Link to comment Share on other sites More sharing options...
Recommended Posts
Please sign in to comment
You will be able to leave a comment after signing in